Answer
Apr 28, 2023 - 04:09 PM
The MSA Altair 5X uses standard MSA 4-gas calibration gas.
The configuration for MSA calibration gas is below
20ppm Hydrogen Sulfide (HS2) 60ppm Carbon Monoxide (CO) 58% LEL Pentane (1.45% Volume Methane as a Simulant) 15.0% Oxygen balance Nitrogen (N2)
Here's a link to the MSA 4-gas calibration gas
The MSA Altair 5X also has exotic sensors that require separate calibration cylinders.
The links for exotic sensor calibration gas is below.
10ppm Chlorine; balance Nitrogen
10ppm Hydrogen Cyanide; balance Nitrogen
10ppm Sulfur Dioxide; balance Nitrogen
10ppm Nitrogen Dioxide; balance Air
The configuration for MSA calibration gas is below
20ppm Hydrogen Sulfide (HS2) 60ppm Carbon Monoxide (CO) 58% LEL Pentane (1.45% Volume Methane as a Simulant) 15.0% Oxygen balance Nitrogen (N2)
Here's a link to the MSA 4-gas calibration gas
The MSA Altair 5X also has exotic sensors that require separate calibration cylinders.
The links for exotic sensor calibration gas is below.
10ppm Chlorine; balance Nitrogen
10ppm Hydrogen Cyanide; balance Nitrogen
10ppm Sulfur Dioxide; balance Nitrogen
10ppm Nitrogen Dioxide; balance Air