Answer
Apr 03, 2023 - 02:20 PM
Here's how to change the detector setting with GCT Manager. This will properly update the calibration set intervals for your MGC-DOCK and all of your MGC Gas Detectors. You can also follow along with the attached video link from Gas Clip Technologies.
- Remove the USB drive from the docking station and insert it into the USB port on your computer.
- Go to GCT manager software and select "USB disk". Make sure to select the proper USB port from your computer downloads page. Once selected click OK.
- After you click OK, GCT Manager will send you to the "Test Station" tab. To the right of the "Test Station" tab will be the "MGC" tab along with any other gas detector models you have downloaded to your USB drive.
- Once the gas detector tab is selected, go to the "Calibration Interval (days)" option and fill in your preferred calibration due date settings.
- When that is completed press the "Write" button near the bottom left of the GCT Manager page.
- Remove your USB drive from the computer and reinsert it back into your docking the station. Once complete click the "calibration" or "bump" button to update the calibration interval onto your docking station.