The Industrial Scientific Ventis Pro5 requires different gas components depending on the sensor configuration. Most Ventis detectors use the following 4-gas configuration:
25ppm Hydrogen Sulfide
100ppm Carbon Monoxide
25% LEL Pentane
18% VOL Oxygen;
balance Nitrogen
The Ventis Pro5 also has exotic sensors that need to be calibrated. Below are the following calibration gas cylinders used to calibrate those sensors
0414 - 25ppm Hydrogen Sulfide, 100ppm Carbon Monoxide, 25% LEL Pentane, 18% VOL Oxygen; balance Nitrogen
0047 - 10ppm Sulfur Dioxide; balance Nitrogen
0073 - 25ppm Nitrogen Dioxide; balance Air
0028 - 10ppm Chlorine; balance Nitrogen
0032 - 10ppm Hydrogen Cyanide; balance Nitrogen
0086 - 2.5% VOL Carbon Dioxide; balance Air
0077 - 1ppm Phosphine ; balance Nitrogen
25ppm Hydrogen Sulfide
100ppm Carbon Monoxide
25% LEL Pentane
18% VOL Oxygen;
balance Nitrogen
The Ventis Pro5 also has exotic sensors that need to be calibrated. Below are the following calibration gas cylinders used to calibrate those sensors
0414 - 25ppm Hydrogen Sulfide, 100ppm Carbon Monoxide, 25% LEL Pentane, 18% VOL Oxygen; balance Nitrogen
0047 - 10ppm Sulfur Dioxide; balance Nitrogen
0073 - 25ppm Nitrogen Dioxide; balance Air
0028 - 10ppm Chlorine; balance Nitrogen
0032 - 10ppm Hydrogen Cyanide; balance Nitrogen
0086 - 2.5% VOL Carbon Dioxide; balance Air
0077 - 1ppm Phosphine ; balance Nitrogen