Apr 28, 2023 - 03:02 PM
The Industrial Scientific iBrid MX6 calibrates to a standard 4-gas configuration and can uses seperate gas blends to calibrate exotic sensors.
Below is the ISC 4-gas calibration gas configuration:
25ppm Hydrogen Sulfide
100ppm Carbon Monoxide
25% LEL Pentane
18% VOL Oxygen
balance Nitrogen
Here is the link for the Calibration Gas Blend 0414 - ISC 4-Gas
Below are the calibration gas blends used for the iBrid MX6 Exotic sensors
50ppm Ammonia; balance Nitrogen
10ppm Chlorine; balance Nitrogen
10ppm Hydrogen Cyanide; balance Nitrogen
100ppm Isobutylene; balance Nitrogen
25ppm Nitric Oxide; balance Nitrogen
25ppm Nitrogen Dioxide; balance Air
1ppm Phosphine; balance Nitrogen
10ppm Sulfur Dioxide; balance Nitrogen
Below is the ISC 4-gas calibration gas configuration:
25ppm Hydrogen Sulfide
100ppm Carbon Monoxide
25% LEL Pentane
18% VOL Oxygen
balance Nitrogen
Here is the link for the Calibration Gas Blend 0414 - ISC 4-Gas
Below are the calibration gas blends used for the iBrid MX6 Exotic sensors
50ppm Ammonia; balance Nitrogen
10ppm Chlorine; balance Nitrogen
10ppm Hydrogen Cyanide; balance Nitrogen
100ppm Isobutylene; balance Nitrogen
25ppm Nitric Oxide; balance Nitrogen
25ppm Nitrogen Dioxide; balance Air
1ppm Phosphine; balance Nitrogen
10ppm Sulfur Dioxide; balance Nitrogen